Introduction
Column chromatography is an extension of thin layer chromatography (TLC). It is assumed that the reader has some experience in laboratory practice and has dealt with thin layer chromatography. Instead of applying a sample on a thin layer of silica or alumina, a sample is deposited on a cylinder of adsorbent and solvent is continually applied with pressure until the components completely drain from the cylinder. With this modification, components can be not only separated but collected into different containers, allowing for purification of mixtures and separate drugs with pollutions. Column chromatography (also known as "flash chromatography"), is frequently used in research settings, as is evidenced by its common occurrence in the procedural sections of journal articles.
Procedural discussion
The same underlying principles of thin layer chromatography (TLC) apply to column chromatography. In fact, a TLC is always run before performing a column to assess the situation and determine the proper solvent ratio. To get good separation, it is ideal if the desired component has a Rf around 0.35 (read about Rf in TLC topic) and is separated from other components by at least 0.2 Rf units. If the spots to be separated are very close (if the difference in Rf is < 0.2), it's best if the middle of the spots has a Rf of 0.35. A Rf near 0.35 is ideal because it is slow enough that stationary-mobile phase equilibration can occur, but fast enough to minimize band widening from diffusion.
There are a few variables which aren't applicable to TLC, but which affect the separation of components in column chromatography. These include the column diameter, quantity of adsorbent used, and solvent flow rate. Pic. 1 show variable recommendations based on the sample size and degree of separation between components. In all scenarios, the columns are to be prepped between 12.5-15 cm high.
There are a few variables which aren't applicable to TLC, but which affect the separation of components in column chromatography. These include the column diameter, quantity of adsorbent used, and solvent flow rate. Pic. 1 show variable recommendations based on the sample size and degree of separation between components. In all scenarios, the columns are to be prepped between 12.5-15 cm high.
For example, a column of 2.5 cm should be able to purify around 400 mg of material if the separation is good (ΔRf > 0.2, third column in Pic. 1), or around 160 mg if the separation is difficult (ΔRf > 0.1). The column should be able to be prepared and eluted using around 200 ml of solvent, and the fractions can be collected with approximately 10 ml of solution each.
There are multiple variations on how to physically run a column. One large difference in methods is how the column is prepared. In the "dry packing" method, dry silica or alumina is added directly to a column, and solvent is allowed to trickle through in portions, then with pressure. In the "wet packing" method, the column is filled with solvent first, then dry silica or alumina is lightly shaken in, then packed with pressure. In the "slurry" method, solvent is added to the silica or alumina in an Erlenmeyer flask, poured onto the column as a sludgy material, then packed with pressure (pressure can be produce by syringe).
It is important to know that heat is liberated when solvent is added to silica or alumina (they have an exothermic heat of solvation). The slurry method is presented in this section, with the main reason being that it allows this exothermic step to happen in an Erlenmeyer flask instead of on the column. If heat is liberated during the packing of the column, it may generate bubbles from the boiling of solvent. These can interfere with the separation of the column if they are not adequately removed, and can crack the adsorbent material in the column.
There are multiple variations on how to physically run a column. One large difference in methods is how the column is prepared. In the "dry packing" method, dry silica or alumina is added directly to a column, and solvent is allowed to trickle through in portions, then with pressure. In the "wet packing" method, the column is filled with solvent first, then dry silica or alumina is lightly shaken in, then packed with pressure. In the "slurry" method, solvent is added to the silica or alumina in an Erlenmeyer flask, poured onto the column as a sludgy material, then packed with pressure (pressure can be produce by syringe).
It is important to know that heat is liberated when solvent is added to silica or alumina (they have an exothermic heat of solvation). The slurry method is presented in this section, with the main reason being that it allows this exothermic step to happen in an Erlenmeyer flask instead of on the column. If heat is liberated during the packing of the column, it may generate bubbles from the boiling of solvent. These can interfere with the separation of the column if they are not adequately removed, and can crack the adsorbent material in the column.
Step-by-Step Procedures
a) TLC of ferrocene/acetylferrocene mixture, b) Column chromatography
The column pictured in this section shows purification of a 0.20 g sample containing a mixture of ferrocene and acetylferrocene (crude TLC is in Fig.2 a). Roughly 8mL fractions were collected into small test tubes, and roughly 400 ml of eluent was used.
a) TLC plate of crude ferrocene/acetylferrocene mixture before elution, b) After elution, c) Column with frit, d) Column with cotton wedged in the bottom, e) Column lacking the cotton wedge (would need to be inserted before use)
Run a TLC
1) Run a TLC on the sample to be purified (Fig.3 a and b) to determine the appropriate solvent for chromatography. The desired component should have a Rf around 0.35 and should ideally be separated from all other spots by at least 0.2 Rf units.
2) Prepare a batch of eluent that gives the proper Rf value. The quantity prepared depends on the quantity of sample, the size of the column, and whether the solvent composition is planned to be changed midway. (See Fig. 1 for guidelines and the eluotropic series for trends in "solvent power").
2) Prepare a batch of eluent that gives the proper Rf value. The quantity prepared depends on the quantity of sample, the size of the column, and whether the solvent composition is planned to be changed midway. (See Fig. 1 for guidelines and the eluotropic series for trends in "solvent power").
Prepare the packed column
3) Obtain an appropriate column (see Fig. 1), and be sure there is something near the stopcock that will allow liquid to pass through, but not solid. Columns may have a sintered disk (also known as a "frit"), Fig.3 c), or a plug of cotton or glass wool remaining from the previous user (Fig.3 d). If no disk or plug is present (Fig.3 e), wedge a small wad of cotton or glass wool into the bottom of the column using a long rod.
a) Pouring in silica to 12.5-15 cm high in the fume hood, b) Pouring the silica into an Erlenmeyer flask, c and d) Making a slurry
Safety note: Powdered silica and alumina are lung irritants, and should always be handled carefully in a fume hood. Spilled dust should be disposed of by mopping it with a wet paper towel (if wet, the fine particles are less dispersive).
5) In the fume hood, pour the adsorbent measured in the column into an Erlenmeyer flask (Fig.4 b), then add some eluent (Fig.4 c). Make a loose slurry by swirling and stirring with a glass stirring rod (Fig.4 d) until all the adsorbent is completely wet, gas bubbles are released, and the consistency is somewhat thick, but pourable.
a) Pouring the slurry into the column, b) Rinsing the flask, c) Adsorbent adhered to the sides of the column, d) Rinsing off the clinging adsorbent
6) Put a beaker or Erlenmeyer flask beneath the clamped column and open the stopcock. In one quick motion, swirl and pour the silica or alumina slurry into the column using a large-mouthed funnel (Fig. 5 a). Immediately use more eluent to rinse residual slurry out of the Erlenmeyer flask (Fig.5 b) and onto the column.
7) Immediately rinse any silica or alumina off the sides of the column reservoir using eluent and a swirling motion from a Pasteur pipette (Fig.5 c and d). If allowed to dry, the adsorbent will cling to the glass and won't easily be rinsed down.
7) Immediately rinse any silica or alumina off the sides of the column reservoir using eluent and a swirling motion from a Pasteur pipette (Fig.5 c and d). If allowed to dry, the adsorbent will cling to the glass and won't easily be rinsed down.
a) Jostling the column to remove air bubbles, b) Applying air pressure, c) Adding sand, d) Rinsing sand from the sides
8) Jostle the column firmly using a cork ring or your knuckles (Fig.6 a) to dislodge any air bubbles in the column (which could cause poor separation or cracking of the adsorbent in the column), and to promote an even deposition of adsorbent.
9) Apply gentle air pressure to the top of the column (Fig.6 b) to compress it, stopping when the eluent level is 1 cm from the top of the column. If a T-adapter is used with the air line as in Fig.6 b, fine control of the airflow can be accomplished by adjustment of the pinch clamp on the rubber tubing. Throughout the entire elution process, keep the white column of adsorbent wet, with the eluent level above the top of the silica or alumina. Gently break the seal to cease application of pressure, and close the stopcock to prevent liquid from dripping out further.
10) Add a thin layer of sand (Fig.6 c), approximately 0.5 cm high. Rinse the sides of the column with eluent using a swirling motion to dislodge the sand off the sides of the glass (Fig.6 d). Open the stopcock and allow liquid to drip out until the liquid is just above the sand layer. Apply air pressure if the dripping is too slow.
9) Apply gentle air pressure to the top of the column (Fig.6 b) to compress it, stopping when the eluent level is 1 cm from the top of the column. If a T-adapter is used with the air line as in Fig.6 b, fine control of the airflow can be accomplished by adjustment of the pinch clamp on the rubber tubing. Throughout the entire elution process, keep the white column of adsorbent wet, with the eluent level above the top of the silica or alumina. Gently break the seal to cease application of pressure, and close the stopcock to prevent liquid from dripping out further.
10) Add a thin layer of sand (Fig.6 c), approximately 0.5 cm high. Rinse the sides of the column with eluent using a swirling motion to dislodge the sand off the sides of the glass (Fig.6 d). Open the stopcock and allow liquid to drip out until the liquid is just above the sand layer. Apply air pressure if the dripping is too slow.
Add the sample
Once the sample is applied to the column, there is a race against time, as diffusion will start to broaden the material. A sample should not be applied until you are ready to complete the column immediately and in its entirety. This process may take between 15 and 90 minutes! If using test tubes to collect fractions, the test tubes should be arranged in a rack before adding the sample, and the column height should be adjusted so that the test tube rack can slide underneath.
a) Dissolving a solid with a small amount of dichloromethane, b) Applying the sample, c) Rinsing the sample in its flask, d) Applying pressure to push the sample onto the column just past the sand layer
11) If the crude sample is a liquid, use it directly (go on to step 13).
12) If the crude sample is a solid, do one of the following things:
12) If the crude sample is a solid, do one of the following things:
a) Ideal situation: dissolve the solid in the minimum amount of the eluent (a few ml at most).
b) If the solid is not particularly soluble, or will not dissolve in a few ml of the eluent, dissolve it in the minimum amount of dichloromethane (few ml at most, Fig.7 a).
c) If the solid is insoluble in the eluent, an alternate procedure is also possible. Dissolve the solid in a round-bottomed flask using a few ml of a low-boiling solvent (e.g. dichloromethane or acetone). Add to the flask approximately 1 g of silica or alumina, then remove the solvent on the rotary evaporator to leave a solid that contains the sample deposited on the adsorbent. With an inch of eluent resting atop the packed column (skip adding the sand layer if this method is used), pour the silica-adsorbed sample onto the column using a wide-mouthed funnel. If any dust clings to the glass, rinse it down with more eluent (go on to step 15).
13) Delicately add the sample to the column via pipette, dripping the liquid or solution directly onto the sand with the pipette tip as close as you can manage, not down the sides (Fig.7 b). Take care to not squirt liquid in forcibly such that indentations would be caused in the sand or silica /alumina column.
14) Rinse the sample container with a little solvent (or dichloromethane if used, Fig.7 c) and add the rinsing to the column using the same pipette (to rinse the pipette as well).
15) Open the stopcock and allow liquid to drip out until the sample is just past the sand layer (Fig.7 d) and into the white area of the column (apply air pressure if this takes more than 20 seconds).
16) Gently rinse the sides of the column with a swirling motion using 1-2 pipettes-full of eluent to rinse any splashed sample. Again, allow liquid to drip out (or apply air pressure) until the sample is pushed into the white adsorbent.
Repeat the rinsing step until you feel confident that the entire sample is deposited on the adsorbent. If some sample is still located in the sand layer, it may dissolve in the eluent when more solvent is added, leading to a loss of yield. If the compound is colored, the rinsing should be completely clear.
14) Rinse the sample container with a little solvent (or dichloromethane if used, Fig.7 c) and add the rinsing to the column using the same pipette (to rinse the pipette as well).
15) Open the stopcock and allow liquid to drip out until the sample is just past the sand layer (Fig.7 d) and into the white area of the column (apply air pressure if this takes more than 20 seconds).
16) Gently rinse the sides of the column with a swirling motion using 1-2 pipettes-full of eluent to rinse any splashed sample. Again, allow liquid to drip out (or apply air pressure) until the sample is pushed into the white adsorbent.
Repeat the rinsing step until you feel confident that the entire sample is deposited on the adsorbent. If some sample is still located in the sand layer, it may dissolve in the eluent when more solvent is added, leading to a loss of yield. If the compound is colored, the rinsing should be completely clear.
a and b) Filling the solvent reservoir, c and d) Eluting the column
Fill with Eluent and Elute the Column
17) Delicately add more eluent via pipette (Fig.8 a), swirling down the sides, and then when the sand layer would no longer be disturbed with the additions, carefully pour in larger quantities (Fig.8 b) of the prepared eluent to fill the reservoir (or fill with as much as may be needed). Clean eluent collected during the packing of the column can be reused.
18) Use air pressure to gently and steadily elute the sample through the column (Fig.8 c and d). The more times the pressure is started and stopped, the more likely the column may crack. It's best if the pressure can be kept gentle and steady the entire time.
The optimal drip rate during elution depends on the size of the column. The ideal flow of eluent is when the solvent in the cylindrical section of the column above the adsorbent drops at 5 cm per minute. Therefore, the drip rate should be slower with a narrow column compared to a wider column. The drip rate for a one-inch column should be where individual drops can be barely distinguished. A stream of liquid pouring out the stopcock with this size of column is slightly too fast.
18) Use air pressure to gently and steadily elute the sample through the column (Fig.8 c and d). The more times the pressure is started and stopped, the more likely the column may crack. It's best if the pressure can be kept gentle and steady the entire time.
The optimal drip rate during elution depends on the size of the column. The ideal flow of eluent is when the solvent in the cylindrical section of the column above the adsorbent drops at 5 cm per minute. Therefore, the drip rate should be slower with a narrow column compared to a wider column. The drip rate for a one-inch column should be where individual drops can be barely distinguished. A stream of liquid pouring out the stopcock with this size of column is slightly too fast.
a-c) Collecting fractions, d) Running a column
Collect fractions
19) Immediately start collecting the eluting liquid into test tubes on a rack (Fig.9 a). See Fig.1 for recommendations on volumes to collect in each test tube.20) When the first test tube fills, or if a certain height of liquid has been collected as recommended by your instructor or Fig.1, move the rack over to start collecting into a different tube (Fig.9 b and c). Fill and keep the tubes in order on the rack.
These different tubes are called "fractions". The goal of a column is to collect small enough fractions that most (or some) fractions contain pure material. If the separation of the mixture is difficult (if the ΔRf of the components is low), it may be best to collect small fractions (e.g. half-filled tubes).
These different tubes are called "fractions". The goal of a column is to collect small enough fractions that most (or some) fractions contain pure material. If the separation of the mixture is difficult (if the ΔRf of the components is low), it may be best to collect small fractions (e.g. half-filled tubes).
a-c) Rinsing material splashed onto the tip of the column, d) Students running a column
21) As liquid drains off the column, it often splashes onto the outsides of the tip of the column, and when the solvent evaporates you may see a ring of material on the tip (you will see a ring of solid if the component is a solid as in Fig.10 b, or oily droplets if the component is a liquid). If the components are colored, the column tip should be rinsed (Fig.10 c) when it appears as if one component has completely eluted and before the other component approaches.
22) Periodically keep an eye on the eluent level, and refill before it drops below the sand layer.
22) Periodically keep an eye on the eluent level, and refill before it drops below the sand layer.
a) Elution, b) Addition of ethyl acetate to increase the solvent polarity, c) Solvent level nears the sand layer, d) Refilling
Possibly Increase the Solvent Polarity
23) One eluent may be used throughout the entire column, especially if the components to be separated have similar Rf values. However, if the components have very different Rf values, the solvent polarity may be increased after one component has eluted from the column (Fig.11 a).
Increasing the solvent polarity will make components travel "faster." There are several reasons a faster elution is desired. First, if one component has already exited the column, the column has already done its job with separation, so speeding up the process will not affect the purity of the collected fractions. Second, the longer it takes to run a column, the wider will be the component bands (due to diffusion), and collecting a broad band of material will use (and waste) a lot of solvent.
Increasing the solvent polarity will make components travel "faster." There are several reasons a faster elution is desired. First, if one component has already exited the column, the column has already done its job with separation, so speeding up the process will not affect the purity of the collected fractions. Second, the longer it takes to run a column, the wider will be the component bands (due to diffusion), and collecting a broad band of material will use (and waste) a lot of solvent.
Contains a partial list of the eluotropic series, a list of common solvents ranked according to their "solvent power" in normal phase chromatography. The more polar solvent causes the most dramatic increase in Rf
24) To increase the solvent polarity, the polar solvent can be dripped directly into the eluent on the column reservoir (Fig.11 b). For example, if using a hexanes:ethyl acetate mixture, addition of pure ethyl acetate to the eluent currently in the reservoir would increase its polarity. If the eluent level is running low, a solution could be prepared that contains a higher percentage of the more polar component. For example, if the column first used a 4:1 hexanes:ethyl acetate mixture, using a 1:1 mixture would be a more polar solvent.
25) Elute the column with the more polar solvent as before, and always remember to watch the eluent level, and refill (Fig.11 d) before it drops below the sand layer.
26) To increase the solvent polarity, the polar solvent can be dripped directly into the eluent on the column reservoir (Fig.11 b). For example, if using a hexanes:ethyl acetate mixture, addition of pure ethyl acetate to the eluent currently in the reservoir would increase its polarity. If the eluent level is running low, a solution could be prepared that contains a higher percentage of the more polar component. For example, if the column first used a 4:1 hexanes:ethyl acetate mixture, using a 1:1 mixture would be a more polar solvent.
27) Elute the column with the more polar solvent as before, and always remember to watch the eluent level, and refill (Fig.11 d) before it drops below the sand layer.
25) Elute the column with the more polar solvent as before, and always remember to watch the eluent level, and refill (Fig.11 d) before it drops below the sand layer.
26) To increase the solvent polarity, the polar solvent can be dripped directly into the eluent on the column reservoir (Fig.11 b). For example, if using a hexanes:ethyl acetate mixture, addition of pure ethyl acetate to the eluent currently in the reservoir would increase its polarity. If the eluent level is running low, a solution could be prepared that contains a higher percentage of the more polar component. For example, if the column first used a 4:1 hexanes:ethyl acetate mixture, using a 1:1 mixture would be a more polar solvent.
27) Elute the column with the more polar solvent as before, and always remember to watch the eluent level, and refill (Fig.11 d) before it drops below the sand layer.
a) Original TLC plate, b) Fractions collected by the column, c) Spotting fractions onto a TLC plate, d) A visualized TLC plate of samples of each fraction
Find and Concentrate the Desired Component
28) In finding the desired component in the test tube fractions, it is helpful to understand the relationship between Rf and elution order in column chromatography.In column chromatography, the sample is deposited leading the column and eluted down, while in thin layer chromatography, the sample is spotted on the bottom of the plate and eluted up. Therefore, a column can be thought of like an upside-down TLC plate. A compound with a higher Rf runs "faster," meaning it will end up higher on a TLC plate, and will be collected first with a column. In the column pictured in this section, the component with the lower Rf (orange on the TLC plate in Fig.13 a), is the component collected second from the column.
29) First determine which test tubes contain dissolved compound.
29) First determine which test tubes contain dissolved compound.
a) Spot a sample of each fraction on a TLC plate labeled with fraction numbers corresponding to the order in which they were collected (Fig.13 c). It may be best to spot each sample 2-3 times over top of one another in case the fractions are dilute.
b) If many fractions have been collected, making you hesitant to sample every fraction, one method to identify colorless fractions that may contain compound is to look for a hint of residue leading the test tubes. After evaporation, a solid residue (Fig.13) or oily droplets are sometimes left on the top of the test tube, making obvious that those fractions contain more than just solvent. Sample all fractions near the tubes containing visible residue.
c) Visualize the spotted TLC plate using UV light and/or a stain to determine which fractions contain compound (Fig.13 d)
a) Eluted TLC plates of fractions containing possible compound, b) Combining fractions, c) Rinsing a fraction tube
28) Run a TLC of all fractions that contain compound, spotting up to five samples per 2.5 cm wide TLC plate. Wider TLC plates may be used for this purpose if available.
29) Identify the compound with the desired Rf by comparison with the original crude TLC plate. Choose to retain the fractions that have the desired compound in pure form, as evidenced by the eluted TLC plate. For example, if the compound with the higher Rf is desired in Fig.15 a, fractions 6-10 should be kept.
30) Combine the pure fractions into an appropriately sized round-bottomed flask (no more than half-full, Fig.15 b). Rinse each test tube with a small amount of eluent (or other solvent if solubility is an issue), and add the rinsing to the round-bottomed flask (Fig.15 c).
31) Evaporate the solvent on the rotary evaporator to leave the purified compound in the flask.
29) Identify the compound with the desired Rf by comparison with the original crude TLC plate. Choose to retain the fractions that have the desired compound in pure form, as evidenced by the eluted TLC plate. For example, if the compound with the higher Rf is desired in Fig.15 a, fractions 6-10 should be kept.
30) Combine the pure fractions into an appropriately sized round-bottomed flask (no more than half-full, Fig.15 b). Rinse each test tube with a small amount of eluent (or other solvent if solubility is an issue), and add the rinsing to the round-bottomed flask (Fig.15 c).
31) Evaporate the solvent on the rotary evaporator to leave the purified compound in the flask.
a) Using air pressure to dry the column, b) Allowing the column to dry upside-down, c) Collecting the waste silica
Clean Up the Column
32) To dry the column, apply air pressure to drain the majority of eluent from the column into a waste container. Then further dry the column using one of these methods:
a) Leave connected a gentle stream of air running through the column to further dry it as you clean up other things (Fig.16 a).
b) Clamp the column upside down over a large waste beaker in the fume hood, so that the adsorbent will fall when it has dried (Fig.16 b). This will take a long time (until the next class period), but is an option.
33) When dry, the adsorbent can be poured out of the column into a waste container in the fume hood (Fig.16 c).
Safety note: Powdered adsorbents are a lung irritant, and their hazard is exacerbated if the column contains residual compounds that can now make their way into your lungs. Pouring silica or alumina powders should always be done in the fume hood.
34) When the majority of the adsorbent has been collected in a waste container, use water to rinse any residual solid into the sink, and then rinse the column with acetone into a waste beaker. Further, clean the column with soap and water and dry with the stopcock parts separated.
Safety note: Powdered adsorbents are a lung irritant, and their hazard is exacerbated if the column contains residual compounds that can now make their way into your lungs. Pouring silica or alumina powders should always be done in the fume hood.
34) When the majority of the adsorbent has been collected in a waste container, use water to rinse any residual solid into the sink, and then rinse the column with acetone into a waste beaker. Further, clean the column with soap and water and dry with the stopcock parts separated.
Troubleshooting
Air Bubbles are seen in the column
An air bubble is an empty pocket where stationary-mobile phase equilibration does not happen, so components move quicker around an air bubble than they should. This may lead to uneven eluting bands, which can cause overlap if the separation of the mixture is difficult (if the components have very close Rf values, as in Fig.17).
If air bubbles are seen in the column and the sand or sample has not yet been applied, give the column a good jostling during packing to remove all air bubbles. See your instructor if the bubbles are not budging, as you may be approaching the task too delicately. If the sand or sample has already been applied, it's best to leave the column as is and hope the air bubbles do not affect the separation.
The Bands are eluting unevenly
If the components of a mixture are colored, it may be obvious when the bands elute in a crooked manner. This is most likely due to the column being clamped at a slight diagonal. If the column is clamped in a slanted manner, components will travel in a slanted manner (Fig. 18). This may cause separation problems if the components have a similar Rf.
There is no way to resolve this problem midway through a column, but if the components have very different Rf values, the slanted bands may have no effect on the separation. In the future, be sure to check that the column is perfectly vertical in both the side-to-side and front-back directions.
Last edited by a moderator: